About the calibration of BME680:
Do you have a table where you have the sensitivity of the sensor (resistance) to every VOC? Or have you calibrated the sensor using a normalized/standardized mix of VOC?
Trying to find out if this sensor is more reactive to some specific gases,(compared with other sensors).
Some other questions:
Is the output static IAQ from BME680, also expressed in a scale from 1 – 500?
The IAQ accuracy reflects the current state of the background calibration process. Does this kind of calibration also apply to static IAQ?
Solved! Go to Solution.
11-21-2020 01:38 AM
BME680 have no selection. so it is sensitive for almost all VOC component.
It is also a relative VOC sensor, not absolute VOC sensor.
static IAQ also have scale from 1 - 500.
The calibration level indicased the overall accuracy from BSEC lib which includes the static IAQ.