11-15-2022 03:42 PM
Temperature works fine but preassure is around 720hpa but should be around 980hpa.
The readouts are
digP1 = 36951
digP2 = -10876
digP3 = 3024
digP4 = 6308
digP5 = 254
digP6 = -7
digP7 = 15500
digP8 = -14600
digP9 = 6000
Adc_P = 522333
11-15-2022 07:04 PM
Hi,
Thanks for your inquiry.
It seems like you are using BMP280. Calculating pressure value needs temperature raw data as well. Please see the attached "sample_calculation_BMP280.xlsx" to check your BMP280 temperature and pressure calculations. You can manually input calibration parameters and ADC_t and ADC_p to see if you can get the correct values. If not, then it means the BMP280 sensor has issue in pressure sensor part. We need you to ship the PCB to your local Bosch office for failure analysis.
Thanks.
11-16-2022 10:23 AM - edited 11-16-2022 01:39 PM
11-16-2022 05:06 PM
Hi,
Thanks for the information.
From your Excel file, the temperature calculation has correct result, while the pressure calculation has wrong result. It seems like both BMP280 and BME280 sensors on the tiny PCBs have issues. If you want, you can ship these two PCBs to your local Bosch support and then we can do failure analysis on these two sensors.
In addition, BMP280 has been end of life. Please switch to BMP390 or the new BMP581 for your projects.
Thanks.