BME280 Calibration values for temperature and pressure giving zero after soft reset

We noticed that sometimes the calculated temperature and pressure measurements showed “0.0” as result. After investigation it was noticed that in these cases the calibration data readout values T1, T2, T3, P1 and P2 were “0” resulting in incorrect temperature and pressure calculation. The value P3 was sometimes a different number than the expected value.

During start-up, the software does configure the BME280 sensor and reads the stored calibration sensor data. Configuration of the sensor is started with reading the Chip ID then sequentially initiate a Soft reset, setting Sleep mode, setting OSR & control values, and then finishing with reading the Calibration values.

To software we used is task/event based, which means that the timing of the sequence with above described functions (or tasks) can be different between start-ups.

Using an oscilloscope, the start-up timing for two situations is measured, one with reading correct calibration data (Figure 1) and one with incorrect calibration data (Figure 2). Logic analyzer signals were added in the code to indicate the start of a task (soft reset, set Sleep mode and reading calibration data). The I2C clock was measured to indicate the communication process.

Figure 1: Correct reading of calibration values

From Figure 1 it can be noticed the time between start of the Soft Reset (SoftRst) and start of reading calibration data (signal RdCal) is ~2.2msec. The blue signal is the I2C SCL clock. We do get the following calibration values:

  • Calibration Data.T1: 28444
  • Calibration Data.T2: 26504
  • Calibration Data.T3: 50
  • Calibration Data.P1: 36562
  • Calibration Data.P2: -10640
  • Calibration Data.P3: 3024
  • Calibration Data.P4: 7491
  • Calibration Data.P5: -157
  • Calibration Data.P6: -7
  • Calibration Data.P7: 9900
  • Calibration Data.P8: -10230
  • Calibration Data.P9: 4285
  • Calibration Data.H1: 75

Figure 2: Incorrect reading of calibration values

From Figure 2 it can be noticed that the timing between starting the Soft Reset (signal SoftRst) and start of reading calibration data (signal RdCal) is much shorter (~1 msec) compared with the timing shown in Figure 1.

This results into incorrect calibration values as shown below.

  • Calibration Data.T1: 0
  • Calibration Data.T2: 0
  • Calibration Data.T3: 0
  • Calibration Data.P1: 0
  • Calibration Data.P2: 0
  • Calibration Data.P3: 2816
  • Calibration Data.P4: 7491
  • Calibration Data.P5: -157
  • Calibration Data.P6: -7
  • Calibration Data.P7: 9900
  • Calibration Data.P8: -10230
  • Calibration Data.P9: 4285
  • Calibration Data.H1: 75

 

According to the BME Sensor datasheet, the start-up time to first communication after power-on is 2ms. Assuming the BME sensor needs some time after Soft Reset initiation (this is not specified in the datasheet), we added an additional delay of 2ms after Soft Reset to prevent reading of incorrect calibration data. 

The question is now, is our assumption right? How much time is needed before reading calibration data after soft reset? Or is this behaviour caused by something else?

 