Hello,
According to the recommendations the BME 688 sensor should be stabilized by running for at least for 24 hours.
For this purpose I'm using HP-354 profile with RDC-1-0 continous readings.
After appr. 30 hours the gas resistance readings seemed to be stable (see attached plot), but rises again later.
How could one explain such a behaviour?
Still needed more time to stabilize ?
Solved! Go to Solution.
11-14-2022 12:00 PM
... performing burn-in process for additional hours (60h) doesn't really helps, RAW gas resistance readings
are still oscillating significantly.
11-22-2022 03:55 AM
Hi fritz,
Take LP mode as an example:
The IAQ value obtained within 8 hours is not credible, and it can be obtained after more than 8 hours。
After 7 days, based on similar baseline, the output value of different sensor in the same environment should be within the scope of the specification, ± 15+± 15%.